There are a couple of settings on the FX-120i that affect the way the scale operates and the accuracy of the result. Stability band width and zero tracking should be zero for the most accurate result. It will also be the most finicky.
Stability band width ( 5t-b ) This item controls the width to regard a weight value as a stable value. When the fluctuation per second is less than the parameter, the balance displays the stabilization indicator and outputs the data. The parameter influences the “Auto print mode”
5t-b 0 This parameter is for sensitive response of the stabilization indicator. Used for exact weighing.
5t-b 2 This parameter ignores slight fluctuation of a weight value. Used to prevent a weight value from drifting due to vibration or drafts.
Zero tracking ( trc ) This function tracks zero point drift caused by changes in the environment and stabilizes the zero point. When the weight data is only a few digits, turn the function off for accurate weighing. Note Digit indicates a unit of minimum weighing value.
trc 0 The tracking function is not used. Used for weighing a very light sample.
trc 1 The tracking function is used. Normal zero tracking.
trc 2 The tracking function is used. Strong zero tracking.
trc 3 The tracking function is used. Very strong zero tracking.
FX120 manual
Stability band width ( 5t-b ) This item controls the width to regard a weight value as a stable value. When the fluctuation per second is less than the parameter, the balance displays the stabilization indicator and outputs the data. The parameter influences the “Auto print mode”
5t-b 0 This parameter is for sensitive response of the stabilization indicator. Used for exact weighing.
5t-b 2 This parameter ignores slight fluctuation of a weight value. Used to prevent a weight value from drifting due to vibration or drafts.
Zero tracking ( trc ) This function tracks zero point drift caused by changes in the environment and stabilizes the zero point. When the weight data is only a few digits, turn the function off for accurate weighing. Note Digit indicates a unit of minimum weighing value.
trc 0 The tracking function is not used. Used for weighing a very light sample.
trc 1 The tracking function is used. Normal zero tracking.
trc 2 The tracking function is used. Strong zero tracking.
trc 3 The tracking function is used. Very strong zero tracking.
FX120 manual