I worked a lot in labs but I have never seen it. All of internal parts are made of metalYes analytical mechanical balances may require grounding to eliminate static charges which will effect accuracy.
I worked a lot in labs but I have never seen it. All of internal parts are made of metalYes analytical mechanical balances may require grounding to eliminate static charges which will effect accuracy.
Is it possible you have a left hand thread on the adjuster (3), this would account for your finding, although I can see no reason why a LH thread would be used."As I understand the operation - turning the adjuster (3) counter clockwise ie. raising it (raising the CofG) will make the balance more sensitive and less stable and will give more deflection for any given weight."
Absolutely agree principally -- I also guessed it. But "facts are stubborn things" -- deflection decreased..
And I don't understand, why..?
Maybe, I will try it again.. but a little bit afraid of making it worse...
Yup! Like he said.What a beautiful old mechanical balance - I've not seen one like that before but do have a couple of old Stanton balances of similar vintage.
As I understand the operation - turning the adjuster (3) counter clockwise ie. raising it (raising the CofG) will make the balance more sensitive and less stable and will give more deflection for any given weight.
One of my Stanton's (AD5)
View attachment 1688673
Great result.Hi 1066!
Now I made a second attempt.
As a starting point I double checked the central position of the libella, and
FIRST I rise the sensitivity rotating screw-3 CCW (as theoretically agreed), It massively shifted zero point (from 0 to approx +5 mg), but the span of deflection was between -5 mg and +5 mg with/without adding 10mg.
THEN zeroing by rotating screw-2 CW (so it moved a bit toward the center) until reaching zero position of the illuminated scale.
Thereafter I tested adding/removing +/- 10 mg -> deflection was 100% OK in both direction.
Thanks for your help and discussion!
![]()
ThanksGreat result.Unfortunately much too sensitive for reloading, unless you are trying to compare kernel size.
Now you need to find another balance in poor condition that you can break up and use the parts to make a superior mechanical beam scale suitable for reloading.