Dear All,
I have a stone-age semi-automatic mechanical analytical balance.
Short description: internal weights can be put by five mechanical dials as 100g/10g/1g/0.1g/0.01g and the last two digits (X.Xmilligrams) can be read from an illuminated marked scale as the moving horizontal rod (scale language).
I compared it with a calibrated digital analytical lab balance and it is basically OK.
This balance can be used a milligram precision weighting basically, but I have two problems with it:
First I tried to shift zero point to the middle of the window by aligning screw 1 and 2 (it was successful)
Then I tried to increase the sensitivity by aligning screw 3.
First surprise: aligning sensitivity shifted also the zero point, too.
Second surprise: originally I supposed that turning screw 3 a bit counterclockwise shifts the center of gravity of the whole moving arm (seesaw) higher so its endpoints will be a little bit wider (up to 10 mg instead of 9). But surprisingly, it happened the opposite: turning screw 3 a bit counterclockwise, it reached 8 mg only.
Thereafter I reset the original positions of both screws as don't want to make the original status worse so the 9/10 of sensitivity still remained.
So any advice/proposal are kindly welcome
)
Thanks in advance,
regards
Peter
I have a stone-age semi-automatic mechanical analytical balance.
Short description: internal weights can be put by five mechanical dials as 100g/10g/1g/0.1g/0.01g and the last two digits (X.Xmilligrams) can be read from an illuminated marked scale as the moving horizontal rod (scale language).
I compared it with a calibrated digital analytical lab balance and it is basically OK.
This balance can be used a milligram precision weighting basically, but I have two problems with it:
- The zero point is not exactly in the middle of the illuminated window and
- Its sensitivity for milligrams is too low. It means that if I (after zeroing) set the last (0.01g) dial from zero to 0.01, the optical scale whows only 9 milligrams instead of 10. In the comparison table it can be clearly seen that the sensitivity bias is higher if the last two digits of comparison weights end 7-8-9 or 10 mg, independently on their absolute weights..
First I tried to shift zero point to the middle of the window by aligning screw 1 and 2 (it was successful)
Then I tried to increase the sensitivity by aligning screw 3.
First surprise: aligning sensitivity shifted also the zero point, too.
Second surprise: originally I supposed that turning screw 3 a bit counterclockwise shifts the center of gravity of the whole moving arm (seesaw) higher so its endpoints will be a little bit wider (up to 10 mg instead of 9). But surprisingly, it happened the opposite: turning screw 3 a bit counterclockwise, it reached 8 mg only.
Thereafter I reset the original positions of both screws as don't want to make the original status worse so the 9/10 of sensitivity still remained.
So any advice/proposal are kindly welcome

Thanks in advance,
regards
Peter