Static effects the lower-priced scales (<$1K), but it is easy to deal with. Electronic labs have to deal with static discharge all the time and there are well-established procedures for dealing with it. Just google ESD, ESD prevention, etc.
I have personal experience with the Acculab VIC-123($350) and the Sartorious GD503($1K). BOTH SCALES worked extremely well for me. But I DO pay attention to proper grounding/static control, surge protection, and creating a stable phyiscal environment(wind, heat, humidity). The GD503 weighs the charge a bit faster and drifts very little...eliminating the need for constant re-zeroing during a session. That said, the VIC-123 was just as reliable if I followed the procedure below.
My VIC-123 procedure:
1. Work in a properly grounded environment: table, equipment, person and make sure the scale has a good quality surge protector.
2. Let the scale warm up for 20-30 minutes
3. Place powder pan #1 on the scale and make sure it is zeroed out and stable. If it is not stable, it either has not warmed up or you have problems with static, wind, etc.
4. Throw a charge into powder pan #2 and then pour this into pan #1 on the scale.
5. Trickle up to final weight directly on the scale. The GD scale settles quite a bit faster than the VIC, but both are accurate.
6. Pour charge into case and place pan #2 back on scale. I use two powder pans because this allows time for the scale to re-zero with pan #2 while I am throwing the next charge with pan #1. This way I can see whether the empty powder pan has re-zeroed before each and every throw.
7. Throw the next charge into pan #2 and repeat the process
8. Once in a while use a check-weight to verify that the scale is spot on after being zeroed repeatedly....it always was.
Sartorius GD503 procedure:
1. Make sure everything is properly grounded, surge protector, etc.
2. Warm up the scale for 15 minutes
3. Calibrate and zero with the single powder pan to be used.
4. Throw the powder, weigh, and trickle up
5. Re-zero every 20-30 minutes if needed: maximum drift over a one hour session is less than .015 grains....usually when I don't let it warm up.
I have heard some folks recommend that you keep your scale always plugged in and turned on. While this may eliminate the need to warm up your scale, I feel this is a BIG mistake. Your expensive scale is exposed to damaging surges in the electrical grid whenever it is connected to said grid. Over time, expensive electronic equipment can be destroyed by the constant barrage of mini-surges and larger surges that occur in most power grids. The best way to protect your equipment is to completely disconnect it from the power source when not in use. Now, the next response people will give is "I'm ok, I have everything plugged into a power strip/surge protector. My response is, "if you want to rely on a surge protector, you better make sure you have a good one that will continue to function over time." MOV-based surge protectors (what most people are using) degrade and fail over time. Check this one out if you want a real surge protector: http://www.brickwall.com/pages/no-failures