I just took delivery on a new A&D 120i scale and I have an AutoTrickler V4 on order. For those of you are using an A&D 120i - do you turn the scale off when not in use or leave it powered up all the time?
 
The manual states warm up ????
I warm my unit up for a couple Hours ,some times over night.
Makes me curious as to why? We’re not making a cake. Is there a circuit board that needs to reach a particular temp of are there calculations going on in there. What’s up ?
 
When you warm up the scale, the electronic components are warming up to their normal operating temperature. At this point the scale will meet the claimed accuracy. The signals, ( voltages and current ) in the electronics are VERY small, and the drift with temperature of the amplifiers, resistors etc, can affect the accuracy until they hit their normal temperatures.
I follow the A&D instruction manual, and let it warm up for 1/2 hour, and turn it off when I am done with it.
 
When you warm up the scale, the electronic components are warming up to their normal operating temperature. At this point the scale will meet the claimed accuracy. The signals, ( voltages and current ) in the electronics are VERY small, and the drift with temperature of the amplifiers, resistors etc, can affect the accuracy until they hit their normal temperatures.
I follow the A&D instruction manual, and let it warm up for 1/2 hour, and turn it off when I am done with it.
Thx for the explanation, I ask questions because I didn’t know the answer, now I do.
J
 
I just took delivery on a new A&D 120i scale and I have an AutoTrickler V4 on order. For those of you are using an A&D 120i - do you turn the scale off when not in use or leave it powered up all the time?
I leave mine on 100% and I’m powering it through a surge protector. I also calibrate it every time I’m going to do a critical work with it.
 
In the olden days of vacuum tubes and other high heat producing electronic components the failure rates of electronics was driven by time in service. Today most electronic and digital equipment is still subject to time in service failure but the time between failures is many times greater. Your computers and iPhone are turned off seldom if ever. The reality is it probably doesn't make much difference.
 
You 100% absolutely need to warm it up.

don’t believe me? if not then try this.

turn scale on, throw a charge on it, record the weight, come back in 30-60 mins and see what the current weight is vs the weight when first turned on.

I do suggest to throw a charge on the pan (so the scale is reading something) while the scale is warming up. not sure why but it seems to help lessen drift. I do like to warm my scale up for at least an hour or two (with a charge on it) to be sure.
 
  • Like
Reactions: MUP
I just took delivery on a new A&D 120i scale and I have an AutoTrickler V4 on order. For those of you are using an A&D 120i - do you turn the scale off when not in use or leave it powered up all the time?
Turning it off doesn't actually turn it off. It sits in a kind of standby state.

The manual is kind of ambiguous on the warmup:

In the section "Assembling":
"Connect the balance to the power for at least 30 minutes with nothing on the weighing pan
before use."

Under the Power Supply section, it reads:
"When the AC adapter is connected, the balance is in the standby mode if the standby indicator
is on (refer to “4. DISPLAY SYMBOLS AND KEY OPERATION”). This is a normal state and
does not harm the balance. For accurate weighing, warm up the balance for at least 30 minutes
before use."

Finally, under "Before Use":
"Plug in the AC adapter and warm up the balance for at least one hour."

Note that it doesn't clarify whether warmup is performed in standby (as implied in the Assembling section) or with the scale actually turned on.
 

Upgrades & Donations

This Forum's expenses are primarily paid by member contributions. You can upgrade your Forum membership in seconds. Gold and Silver members get unlimited FREE classifieds for one year. Gold members can upload custom avatars.


Click Upgrade Membership Button ABOVE to get Gold or Silver Status.

You can also donate any amount, large or small, with the button below. Include your Forum Name in the PayPal Notes field.


To DONATE by CHECK, or make a recurring donation, CLICK HERE to learn how.

Forum statistics

Threads
166,310
Messages
2,216,092
Members
79,543
Latest member
drzaous
Back
Top